I've recently connected a network of 14 thermistors in parallel to a labjack t7 (outfitted with mux80 and 4x cb37 expansion boards). I used excitation circuit #4 - and confirmed voltage readings (and corresponding resistance/temp calculations). All 14 measurements are solid. However, when I introduce 3 differential voltage measurements (each measuring a battery voltage), all thermistor voltages drop significantly. My understanding was that by measuring the batteries as differential pairs, they are isolated from single-ended thermistor measurements. Any ideas as to why the battery voltage measurements are interacting with the thermistor voltages?
Unfortunately, true isolated measurements are not possible on the T7 without external hardware. Your battery measurements as they are do not look valid; you must have a DC path to GND at the AIN inputs for bias current to flow. The following section has a related discussion. I would also recommend seeing the section about measuring power supplies in series:
http://labjack.com/support/app-notes/differential-analog-inputs#differen...
Another part of the issue might be the source impedance of the thermistor circuits. You could see channel settling issues with source impedance above 1kΩ. I would recommend seeing our settling time application note:
https://labjack.com/support/app-notes/SettlingTime