Hello,
We are attempting to comply with certain safety specifications where one of the topics is the calibration of measuring equipment. I have a few questions for the Labjack team:
Can Labjack device be calibrated and how? (we mainly use T4 and T7 devices)
Do you have any customers that actually perform the calibration?
Do you recommend any calibration procedure, calibration intervals?
Thank you for looking into this!
Mina
Hello, I had answered my questions. Appologies for not checking this earlier.
I can now see that you offer the calibration service with certificate, https://labjack.com/products/calibration-service-cert-u6ue9t7
No need to answer to my previous questions. Instead I would like to ask this:
You mentioned that you do not do any calibration for T4 devices. If some sort of compliance is needed for this devices, do you recommend to customers to simply develop the calibration procedure themselves?
Yes, that would be our recommendation. The "Is My Calibration Good" section suggests how to do this:
https://labjack.com/products/calibration-service-cert-u6ue9t7
If you attempted this verification and found that new cal constants are needed to be in spec, you could write new values or send the unit to us:
https://labjack.com/support/datasheets/t-series/internal-flash/t4-calibration-constants