Dhananjayan
dhananjayan's picture
COUNTER INPUT 16,17 NOT WORKING ON T7 PRO

Hi im using t7 pro and labview for programming. i tried to run on your example program stream mode program

when i try to run the program the following error occurred .

need your support for solving this problem.

File Attachment: 
LabJack Support
labjack support's picture
That example enables CounterA

That example enables CounterA and CounterB, but the problem is those counters share resources with Clock0, and Clock0 is enabled by default.  See fixed example attached which does writes to make sure Clock0/1/2 are all disabled.

https://labjack.com/support/datasheets/t-series/digital-io/extended-features/high-speed-counter

 

File Attachment: 
Dhananjayan
dhananjayan's picture
Thank you so much!

Thank you so much!