Frecuency Inputs | LabJack
 

Frecuency Inputs

2 posts / 0 new
Last post
JUAN EFFIO
JUAN EFFIO's picture
Frecuency Inputs

I need 5 frecuency inputs, that is possible? I am working with a T7

LabJack Support
labjack support's picture
You sent an email also and

You sent an email also and told us the max frequency of these signals is only 10 Hz.  That means the "Interrupt Frequency In" feature, of which there are 6 available, is a likely solution for you.  See Table 13.2-2 at the end of this page:

https://labjack.com/support/datasheets/t-series/digital-io/extended-feat...

You could also consider counting pulses rather than timing pulses:

https://forums.labjack.com/index.php?showtopic=1105

The T7 has 6 interrupt counters plus 4 hardware counters, and in fact we have a Lua script example that implements 107 counters using the 23 DIO and 84 AIN (with Mux80):

https://labjack.com/support/software/examples/lua-scripting/counter-exam...