When I start LJstreamM I get an error LabJack Error #2621: STREAM_SETTLING_INVALID. I saw a post were to set STREAM_SETTLING_US to 1. Which then the error changed to LJME_NO_RESPONSE_BYTES_RECEIVED 1263.
Set STREAM_SETTLING_US = 0 for auto settling. If you still get the error please send a screenshot showing all your LJStreamM settings and the error message:
Here is the setting that we used for our project. We have the load cells unplugged after the screenshots are taken but the error is the same as they were plugged in. The third sensor (rotary encoder) is still plugged in while the screenshots are taken.
I see that you are doing 6 channels at 500 scans/second which is 3000 samples/second, and that your smallest range used is 0.01V. If you go to Table A.1.7 you can see that multi-channel streaming is not supported at Range = 0.01V, and even if you make your minimum range 0.1V you would still be exceeding the 1.7k max sample rate listed there.
Set STREAM_SETTLING_US = 0 for auto settling. If you still get the error please send a screenshot showing all your LJStreamM settings and the error message:
https://labjack.com/support/app-notes/screenshots
Here is the setting that we used for our project. We have the load cells unplugged after the screenshots are taken but the error is the same as they were plugged in. The third sensor (rotary encoder) is still plugged in while the screenshots are taken.
Thank you for trying to help us.
I see that you are doing 6 channels at 500 scans/second which is 3000 samples/second, and that your smallest range used is 0.01V. If you go to Table A.1.7 you can see that multi-channel streaming is not supported at Range = 0.01V, and even if you make your minimum range 0.1V you would still be exceeding the 1.7k max sample rate listed there.
https://labjack.com/support/datasheets/t-series/appendix-a-1