LJME errors

4 posts / 0 new
Last post
Abosaad
abosaad's picture
LJME errors

Hello,

When I start LJstreamM I get an error LabJack Error #2621:  STREAM_SETTLING_INVALID. I saw a post were to set STREAM_SETTLING_US to 1. Which then the error changed to LJME_NO_RESPONSE_BYTES_RECEIVED 1263. 

Thank you,

LabJack Support
labjack support's picture
Set STREAM_SETTLING_US = 0

Set STREAM_SETTLING_US = 0 for auto settling.  If you still get the error please send a screenshot showing all your LJStreamM settings and the error message:

https://labjack.com/support/app-notes/screenshots

 

Abosaad
abosaad's picture
Here is the setting that we

Here is the setting that we used for our project. We have the load cells unplugged after the screenshots are taken but the error is the same as they were plugged in. The third sensor (rotary encoder) is still plugged in while the screenshots are taken.

Thank you for trying to help us.

File Attachment: 
LabJack Support
labjack support's picture
I see that you are doing 6

I see that you are doing 6 channels at 500 scans/second which is 3000 samples/second, and that your smallest range used is 0.01V.  If you go to Table A.1.7 you can see that multi-channel streaming is not supported at Range = 0.01V, and even if you make your minimum range 0.1V you would still be exceeding the 1.7k max sample rate listed there.

https://labjack.com/support/datasheets/t-series/appendix-a-1