Strain Gauge Measurement with T7

I just have a few questions about the best way to set up a system to measure 4-350 ohm strain gauges at the same time.  Will I need to link 1 LJTick-InAmp for each bridge circuit?  Will the T7 be able to provide the excitation voltage needed to run 4 bridge circuits?

LabJack Support
labjack support's picture
The InAmp has two differntial

The InAmp has two differntial channels. So, one InAmp can be used to condition two bridge circuits. Each InAmp has a Vref that can be used to power the bridges in parallel. I would power the bridge from the InAmp that measures it. That would be two bridges powered by each InAmp.

Tony Neufeld
tonyjneufeld's picture
Would it be possible to run

Would it be possible to run the strain gauges without the InAmps?

LabJack Support
labjack support's picture
The T7 can read the strain

The T7 can read the strain gauges directly. I recommend using differential mode, so you will probably need the CB37 to get to the 8 AINs you need. The Vs terminals are too noisy to be used to power the bridge circuits. The DACs may be OK, but a good reference supply would be best.

Tony Neufeld
tonyjneufeld's picture
Do you have any suggestions

Do you have any suggestions for the reference supply?

LabJack Support
labjack support's picture
Check out item #1 of post #13

Check out item #1 of post #13 here, and let us know if you need more help:

https://forums.labjack.com/index.php?showtopic=4376