I'm trying to set the default values for my U3-LV. All the FIO Digital State are uncheck (low). But whenever I press the Write Values buttons, all the Digital State FIO channels are saved as high. Is that normal?

UD Driver Version: 3.48 - Up To Date

U3-LV serial: 320059035

LJ Control Panel: Ver 2.33

LabJack Support
labjack support's picture
That is normal when the

That is normal when the Digital Direction is configured for input (unchecked). In this case the state is not configurable, and they default to high when reading the settings from the firmware.

If the Digital Direction is an output (checked), the state is configurable and its setting is retained after "Write Values" is clicked.