In LJLogUD there is a place to specify the number of channels being used, and it goes up past 16. There are 16 rows in the main panel. Is there a way to view more than 16 rows worth of channels? I want to use all 16 of my possible digital input (FIO4-7, EIO0-7, and CIO0-4) and read some analog signals using FIO0-3 at the same time. So if I set the number of chnnels to 20, how can I see all 20 "rows" worth of channels. I am using the 16 already there by default for the digital input acquistion.
Thank you.
You only need 2 rows (with channels 193 & 194) to read and log all digital I/O, but if you want to apply scaling formulas such that you see each individually as a 1 or 0 in real-time then each will use 1 row for each. 16 rows is a limitation of the simple LJLogUD software.
https://labjack.com/support/u3-software-options
Thank you, that answers my question.